iNexiv Rotary Indexer Setup and Calibration

Software topics related to Nikon iNexiv and Retrofit Camera
Post Reply
User avatar
US_Helpdesk
Moderator
Moderator
Posts: 1087
Joined: Wed Feb 23, 2011 7:26 pm

iNexiv Rotary Indexer Setup and Calibration

Post by US_Helpdesk » Tue Sep 15, 2015 4:38 pm

1) Enable the Rotary Indexer from Machine Setup

From System>Setup>Machine Setup choose the Rotary Table tab.

Check the Enable Rotary Table check-box

Table Type -> Other Add-on Rotary Table

Direction of rotary table normal -> normally X+ on VMA-2520 (could be any orientation except Z-)

Select DCC option

Controller Interface -> Newmark NSC-A1

Select Indexing

Encoder Type -> Add-on
roatry_machine_setup.jpg
2) Calibrate Rotary Table / Indexer

From System>Setup>Rotary Table choose Calibrate Table
rotary.jpg
Set Calibration Parameters as shown below:
rotary_iNexiv.jpg
Note the large Clearance Distance. This prevents collision of artifact when indexing during calibration routine.

Note the actual sphere and stem diameter and direction for the calibration artifact. We suggest something like this:
rotary_image.jpg
Click OK and follow all prompts. Software will prompt user to place probe above sphere artifact for first two positions, all subsequent positions will execute in full DCC. Calibration will execute twice and then report overall radial deviation. If this deviation is acceptable click OK to complete.

Note: Typical radial deviation is usually a couple microns (0.002mm)

32-bit and 64-bit drivers are here:
32_bit_Newmark_Drivers.zip
64_bit_Newmark_Drivers.zip
You do not have the required permissions to view the files attached to this post.
I've migrated to a new user account, see my other posts here

Tabor
Super User
Super User
Posts: 175
Joined: Thu Dec 20, 2012 3:43 pm

Re: iNexiv Rotary Indexer Setup and Calibration

Post by Tabor » Wed Sep 16, 2015 10:49 am

Just for clarity: All the settings are in MM, correct?
Tabor

User avatar
US_Helpdesk
Moderator
Moderator
Posts: 1087
Joined: Wed Feb 23, 2011 7:26 pm

Re: iNexiv Rotary Indexer Setup and Calibration

Post by US_Helpdesk » Thu Sep 17, 2015 12:51 pm

All linear units are in your current units. Could be Inch or mm depending on current settings. In my case the units were mm's.
I've migrated to a new user account, see my other posts here

User avatar
CrashN8
Site Admin
Site Admin
Posts: 520
Joined: Thu Feb 09, 2017 4:07 pm

Re: iNexiv Rotary Indexer Setup and Calibration

Post by CrashN8 » Wed Sep 06, 2017 12:06 pm

Here are some details on a suitable calibration artifact - notice I've used M3 components due to added rigidity compared to M2 counterparts:
renishaw_rotary_calibration_artifact.jpg
Here are the links to each component:

http://www.renishaw.com/shop/Product.as ... -5000-7610
http://www.renishaw.com/shop/Product.as ... -5004-7584
http://www.renishaw.com/shop/Product.as ... -5000-7630
You do not have the required permissions to view the files attached to this post.